Thread: ADC value Units

  1. Default Re: New USB Thoughtstream, new software?!

    @ Andy:

    So that means each measurements that is displayed in the interface software is really the average of 6 measurements? And these exact measurements are still saved in the txt. file individually?

    Thanks,
    Thomas
    Last edited by Andy; 05-05-2011 at 02:14 PM. Reason: clarification of which TS

  2. #12
    Join Date
    Apr 2006
    Location
    Calgary, Alberta, Canada
    Posts
    4,117
    Blog Entries
    3

    Default Re: New USB Thoughtstream, new software?!

    Quote Originally Posted by Thomas
    So that means each measurements that is displayed in the interface software is really the average of 6 measurements?
    Yes.
    Quote Originally Posted by Thomas
    And these exact measurements are still saved in the txt. file individually?
    No. The averaged samples are saved.

    There's a fix in the latest version of the interface software for the TS USB where now the timestamp starts when you start the session. Instead of an absolute time-of-day, it's in milliseconds from the time the session was started. The 1st entry is at 0.30 seconds.

    The Original interface software for the RS-232 Thoughtstream writes to the file 2 times/second but does not write a timestamp.

    -Andy.
    Last edited by Andy; 05-05-2011 at 02:22 PM. Reason: clarification

  3. #13
    Join Date
    Apr 2006
    Location
    Calgary, Alberta, Canada
    Posts
    4,117
    Blog Entries
    3

    Default Re: ADC value Units

    Quote Originally Posted by insk View Post
    Could it be, that my thoughtstream device is not the USB version, because I connect it with the PC by using a serial-to-usb adapter and not a common USB cable!
    That is correct. The USB version has a USB connection on the console. You have the RS-232 version.
    -Andy.

    Hey, if someone makes a good post, don't forget to click at the bottom of their post to add to their reputation!

  4. #14

    Default Re: ADC value Units

    great. thanks. but i still don't know or can't figure out, how often the probe is written to the Txt file!

    here a short excerpt of the txt file that has been produced by the thoughtsream:
    1379
    1382
    1383
    1383
    1382
    1382
    1382
    1384
    1388
    1349
    1303
    1246
    1176
    1091
    1032
    ...

    thank you

  5. #15
    Join Date
    Apr 2006
    Location
    Calgary, Alberta, Canada
    Posts
    4,117
    Blog Entries
    3

    Default Re: ADC value Units

    Ok, here's the confirmed info:

    On the Original Thoughtstream (RS-232) the data sent out the RS-232 port and written to the .txt file is at 2 datum/sec or every 0.5 seconds.
    The data output to the text file is only the ADC value.
    To convert to probe resistance, use the forumula:
    Probe Resistance (in MegOhms) = (1925 / (12 bit ADC value)) - 0.47

    On the Thoughtstream USB the data sent out the USB port and written to the .txt file is at 3.33 datum/sec or every 0.3 seconds.
    The data output is the ADC value and a timestamp.
    To convert to probe resistance, use the formula:
    Probe Resistance (in MegOhms) = (7700 / (14 bit ADC value)) - 0.47
    Last edited by Andy; 05-15-2011 at 07:41 AM. Reason: Updated info
    -Andy.

    Hey, if someone makes a good post, don't forget to click at the bottom of their post to add to their reputation!

  6. #16

    Default Re: ADC value Units

    Thank you very much for your help!
    Last edited by Andy; 05-15-2011 at 07:41 AM.